Feature

General describe:

USO-2300 is a digital storage intelligent oscilloscope with high performance.Except the oscilloscope general features,there are some special and useful function in our USO-2300 oscilloscope.

  • Plug and play.Small Size and Metal cabinet.
  • USB power supply directly. No external power supply required.
  • Multi-functions: FFT freqency Spectrum Scope,wave recorder,signal generator,freqency counter,voltage metre,stero viewer,correlation viewer and digital storeage intelligent oscilloscope.
  • Record and play ."see and record and playback",the computer can not only save current wave view on the screen, but also record continuous wave. The operation of recording and playback , is just as to operate DVD.
  • The recorded data can be the format that the application Excel,Word can use.
  • The Wave view can be expanded simply.the wave view can be saved as jpg file.
  • Frequency Spectrum Scope (FFT) / Lissajous pattern (X-Y PLOT)/S MAP/ C MAP.
  • The variables of input signal be calculated and displayed.User can adjust the displayed location and color.All variables can be copied to Excel,Word,etc.
  • Auto calibration on-line.
  • User can adjust triggler level and triggler horizontal position.
  • Print on-line.
  • Menu style and 3D style for knobs can be selected.
  • Marker,hot key,color setting and graphic operation .
  • Intelligence frequency search and trace.
  • Hardcopy Function: Simple "Copy" and "Paste" into other applications such as Word, Excel.
  • Channel mathematic operations.
  • User can save and restore the settings.
  • One To More. Several USB Digital Oscilloscopes can be connected with one computer and run correctly.More channels can be measured by one computer.
  • Inherit the advantage of the traditional instrument ,Friendly man-machine interface ,Easy to operate.

USB Oscilloscope Index:

  • Voltage division : from 10mV/Div to 5V/div ,total 9 Divisions.
  • Timebase: from 1S/Div to 20ns/Div,total 24 divisions.(20ns and 50ns is forbidden in Low frequency model)
  • Coupled: AC/DC/GND.
  • Triggle: 3 triggle styles.
  • Input resistance:1Mhom.
  • Input capacitance:18pF.
  • Maximum input voltage: 50Vpk+DC.
  • Working temperature: 0°C~40°C.
  • Storage temperature: -40°C~70°C.
  • Recorder capacity: 500MB.
  • Protocol: Compatible with USB 1.1/2.0.
  • Compensation signal frequency: ~0.5V@1KHz.
  • Noise: 2 LSB.

OS:

  • WINDOWS 98
  • WINDOWS 2000
  • WINDOWS XP

Installation And Uninstall

USO_2300 Application:

  • USO_2300 Application no need installation. Simple to download and extract,it can be run.
  • If the directory the software in is deleted,the software is deleted at all.
  • The application is USO_2300.EXE.

Device Driver:

  • While the USB oscilloscope is used at the first time,the computer will prompt to install device driver .
  • Before install device driver,Pls temporarily disconnect your PC from the Internet.
  • The device driver is the file USO.INF under the directory \DRIVER.
  • the device driver uninstaller is the .exe file under the directory \DRIVER.
  • User can uninstall the device driver through the windows "CONTROL PANEL"
  • The USB oscilloscope device is named as "RuiQi USO Device" in the computer system hardware list.

USB Oscilloscope (96138 USB Oscilloscope):

  • Small Size and Metal cabinet.
  • USB power supply directly. No other power supply. Plug and play.
  • The application can automatically search USB oscilloscope and detect the device model.
  • If the application find a USB oscilloscope,the model of the device and the application startup number will be displayed at the top of the main window.The product ID will be displayed at "About" window.
  • One computer can be connected with several USO-2300 oscilloscopes.They differ in the application startup number and the model.

Config Requirement:

  • If the application can be run at the DEMO mode,the computer can drive one or more USB oscilloscope.
  • Monitor resolution: at least 800*600. The best resolution: 1024*768.
  • RAM :at least 128M.

How To Use

1. Operation Describe :

A. Click

  • Operate all menu items and buttons .
  • Set TIMEBASE ,VOLT/DIV ,POSITION and TRIG LEVEL.
  • Change color or switch option.
  • Move wave view of time domain to left or to right.

B. Double Click

  • Double Click measure view,the measure setting window will appear.

  • We can change the displayed content and color.

C. Right click

  • At wave view,a new menu will appear.
  • Copy ,Print,Setting ,Close...will appear.

D. Drag

  • Move the marker in the time domain wave view or frequency domain view.
  • Move the triggle indicators to adjust the triggle level and triggle horizontal position.
  • Operate TIMEBASE ,VOLT/DIV ,POSITION and TRIG LEVEL.
  • Change the size of the time domain wave view or frequency domain view.
  • Move wave view of time domain to left or to right.

E. Default Abort and OK button on the Setting window

  • Default: Restore factory setting.
  • Abort :Close the setting window,no save.
  • OK:save and close the setting window.

F. Hot key

  • Hot key is the letter with underline.

G. Auto calibration

  • Before use the device at the first time,the device should be calibrate ,and the calibration result should be saved.
  • Run auto calibration,user should operate according to the application prompt.
  • Abort the calibration procedure or after calibration,the device should be unplag or the application should be restart once.

H. Selected

  • If Selected,the font or the line will turn to colored.otherwise,it appears black or grey.

2. Menu Items:

A. Main window FILE menu

  • FILE | OPEN Open the saved file,to display the saved wave data
  • FILE | SAVE Save current wave data
  • FILE | PRINT SETUP
  • FILE | PRINT
  • FILE | BATCH DELETE Delete USO data file
  • FILE | EXIT

B. Main window VIEW menu

  • VIEW | FFT MAP (Frequency Spectrum Scope)
  • VIEW | S MAP (Stero Map)
  • VIEW | C MAP (Correlation Map)
  • VIEW | X-Y PLOT (Lissajous pattern)
  • VIEW | CURSOR
  • VIEW | MENU SWITCH
  • VIEW | TOOL BAR
  • VIEW | MENU SWITCH
  • VIEW | Maximize Main View
  • VIEW | Minimize Control Panel
  • VIEW | Normalize Control Panel
  • VIEW | Minimize Measurement Panel
  • VIEW | Normalize Measurement Panel

C. Main window SETUP menu

  • SETUP | PROBE
  • SETUP | MAIN WINDOW
  • SETUP | FFT WINDOW
  • SETUP | X-Y PLOT
  • SETUP | PLAYBACK
  • SETUP | LANGUAGE
  • SETUP | AUTO CALIBRATE (Enter auto calibration procedure.)

D. Main window RECORDER menu

Record As USO File: Record as the USO file that can be playback by USB scope application.

Record As TXT File: Record as the file that the application Excel,Word ,Notepad can use.

  • Record. Enter RECORDER mode,the recorder bar and the recorder time bar will appear.
  • Buttons are:Pause | Exit record| start to record.
  • When record,time bar turns to colored and count.
  • Record will be correct and no repeat .

E. Main window PLAYBACK menu

  • Playback. Enter PLAYBACK mode,the playback bar and the progress bar will appear.
  • Buttons are:Playback| Pause|Exit PLAYBACK|Next frame|previous frame|return the first frame.
  • Playback progress bar,display the playback location.while pause,user can change the playback loction.
  • The playback speed can set at SETUP| PLAYBACK.

F. Main window A+/-B menu

  • switch A+B ,A-B and close A+/-B.

G. Main window Cursor,FFT,S Map,C Map,X-Y Plot menus

  • switch the cursor(marker),FFT, S Map,C Map,X-Y Plot.

H. Main window MSave menu

  • Mode save.Save all the current setting as a file.

I. Main window MRestore menu

  • Mode Restore.Restore the setting file.

J. Main window HELP menu

  • HELP | User Manual
  • HELP | Homepage
  • HELP | About

3.Buttons

A. AUTO

  • Enter AUTO mode.

B. NORMAL

  • Enter NORMAL mode.

C. SINGLE

  • Enter SINGLE mode.

D. HOLD

  • Enter HOLD mode.

E. PLAYBACK

  • Enter PLAYBACK mode.

F. DEMO

  • Enter DEMO mode.

G. DC

  • DC coupled.

H. GND

  • Display GND location.

I. AC

  • AC coupled.

J. TRIG A and TRIG B

  • Select the triggle source:Channel A or Channel B.

K. TRIG mode

  • Select triggle mode:positive edge ,pulse or negative edge.

L. FA

  • Frequency search and trace at channel A.

M. FB

  • Frequency search and trace at channel B.

N. NFT

  • Close frequency search and trace function.

O. Reset Buttons

  • Click the reset buttons,can reset the voltage level.

P. Up/Down Buttons

  • Click the up-down buttons,can adjust the voltage level.

4.Knobs

A.TIMEBASE

  • Click or drag can change timebase.

B. VOLT/DIV

  • Click or drag can change channel voltage devision.
  • The list box can change voltage devision directly.

E. POSITION

  • Click or drag can change the vertical position of the channel A,chanel B and triggle level.
  • The list box can change the vertical position of the channel A,chanel B and triggle level directly.
  • The up/down keys can also adjust the vertical position of the channel A,chanel B and triggle level directly.

 

5.Suggestion:

  • If the voltage of the input signal to BNC is not more than 2 times of the current voltage division full scale,more precis waveform will be captured.For an example,the 2 times of the 1V/Div full scale is 20V.
  • Almost probes at 1X (1:1) limit the bandwidth of the instrument to 6MHz.To use the full bandwidth of the oscilloscope,the probe should switch to 10X(1:10).The high-end probe(such as Tektronix P2200) will improve the instrument performance.

6.Warning:

  • Because the metal cabinet of the USB oscilloscope is connected with the PC ground,the desktop computer customers should use the electrical outlet with 3 terminals.
  • To void harm to the oscilloscope and the probes,don't put them into the fog,liquid and menstruum,don't put them into the environment with conductive pollution.

Probe Capacitor Compensation :

  • Before use probe at 1:10 or 1:100,the compensation capacitor of probe should be adjusted.
  • Set the timebase to 1ms/div.Connect probe terminal with 1KHz compensation terminal.Connect the probe ground with the oscilloscope metal cabinet.
  • To compensate the probe at 1:1,set the voltage devision to 0.5V/div.To compensate the capacitor of the probe at 1:10 or 1:100,set the voltage devision to 0.05V/div.
  • Adjust the compensation capacitor of probe,until 1KHz wave show the normal square wave.

Others:

Copyright:

  • All right reserved.
  • If you use the product,you must promise not to to disassemble the product(include software and hardware).

PC software and driver upgrade :

  • No Charge.
  • Update or upgrade will be noticed at our website.

Hardware (include firmware) upgrade:

  • One series Product can be upgraded to the high end model commonly.
  • Before user decide to upgrade hardware,Pls give us an Email to confirm it.

Web site homepage:


[该贴于2006-6-1 21:51:13被admin编辑过]


发贴时间:2006-5-24 18:46:54IP:219.133.*.*
√ 快速回复 
回复主题 USO-2300 User manual(English)
输入用户名和密码: 用户名: 注册  密码: 忘记密码
内容



支持3种图片发布方法:

1.您可以直接上传本地图片。

2.您可以贴粘图片网址然后点击“插入图片URL”按钮,将网址贴粘后确定。

3.您可以从别的网站的网页上直接将内容拷贝后贴粘到右边的输入框中。

注意:
1.本社区用户请确认您发表的言论符合《互联网电子公告服务管理规定》、《全国人大常委会关于维护互联网安全的规定》,并符合本论坛的主旨
2.本社区用户不得在本社区发表包含破坏宪法和法律、破坏民族团结、捏造或者歪曲事实、宣扬封建迷信、淫秽、色情及其他违反宪法和法律行政法规的言论。
3.本社区用户因为违反本社区规定而触犯中华人民共和国法律的,责任自负,本网站不承担任何责任。

发贴内容并不一定代表本站观点!版权归原作者,其它媒体或网站转载请与本站或原作者联系,并注明出处。
96138.com 瑞棋科技 版权所有